X-Ray Forum for NDT Users
It is currently 16.04.2024, 22:57

All times are UTC + 1 hour




Post new topic Reply to topic  [ 2 posts ] 
Author Message
PostPosted: 09.12.2019, 17:54 
Offline
User avatar

Joined: 19.10.2019, 15:47
Posts: 135
The scanning method is the master method - still.
In front of the tube is a cross table with micrometer stepper, which moves a pinhole of about 10µm*10µm size across the area of the beam.

The picture shows the system (without the computer) in front of a 160kV X-ray tube.


The double slit collimator is made from four blocks of tungsten of 400x200x200mm^3 each, which have four 10µm thin 190x190mm^2 foils of steel (red color) to form in the center a hole of 10µm*10µm (best visible in Top View):


The collimator is moved step by step in both directions in front of the tube; this procedure used to take about one hours per focal spot. In the end an isodose contour plot was created.
Here is as example from a Philips MCN 451 X-ray tube the small focal spot:

The red X marks the maximum intensity (100% value); the following colored lines mark 10% less dose as further the line is away from the center; e.g. the outer red line is the 20% isodose contour line. The 10% threshold line to the background signal determines the nominal size which is here 1.487 x 1.448 mm^2 (width x length).

The Philips MCN 451 X-ray tube has two focal spots, the isodose contour plot of the large spot shows higher values:

and comes out with a focal spot size of 5.591 x 7.259 mm^2.

EN12543-1 recommends values for focal spot classes in [mm]:

and with this values the small focal spot of the MCN451 was nominated to 1.8mm and the large focal spot to 8mm.

Finally there is a bad news: There is no more equipment which can do this procedure; the last set of equipment at YXLON was scrapped in 2013 as there were no more spareparts available for the old evaluation PC. As far as I know it is the same with the equipment at BAM and also the Seifert/GE system shall be not more operable. In the conclusion this part shall be withdrawn in the next revision of EN12543 and Part 2 shall become the master.


Top
 Profile  
Reply with quote  
PostPosted: 13.12.2019, 12:29 
Offline
User avatar

Joined: 19.10.2019, 15:47
Posts: 135
Here is - for reference - the pinhole film-image of the small focal spot (the crossed wires were used to determine the center of the focal spot)


and here the pinhole film-image of the large focal spot (the crossed wires were used to determine the center of the focal spot)


Top
 Profile  
Reply with quote  
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 2 posts ] 

All times are UTC + 1 hour


Who is online

No registered users


Similar topics

EN 12543-4 Part 4: Edge method
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0
EN 12543-3 Part 3:Slit camera radiographic method
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0
EN 12543-5 Part 5: For mini- and µ-focus X-ray tubes
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0
EN 12543-2 Part 2: Pinhole camera radiographic method
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0

You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum

Search for:
Jump to:  
cron



Hosted by iphpbb3.com

Impressum | Datenschutz