Part 3 of the EN focal spot measurement standard uses a similar method as the pinhole evaluation method in part 2. But instead of a round pinhole a slit is taken as aparatus and two separate measurements have to be done.

1 - focal spot
2 - direction of the beam
3 - maximal deviation of the symmetrical axis of the slit

The larger of both values has to be taken as focal spot value. The user of this standard has to prove, that the deviation to part 1 is smaller than +/-10%.
The slit has an opening size of 10µm with +/-5µm tolerance. Similar to part 2 there is a table for a minimum magnification - but this table starts at 100µm ...

As far as I know this part of the standard is not very often used. Also I did not found an updated version which would allow a digital image capture and evaluation like part 2 got in the year 2008.


Top
 Profile  
Reply with quote  
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 1 post ] 

All times are UTC + 1 hour


Who is online

No registered users


Similar topics

EN 12543-5 Part 5: For mini- and µ-focus X-ray tubes
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0
EN 12543-2 Part 2: Pinhole camera radiographic method
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0
EN 12543-1 Part 1: Scanning method
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 1
EN 12543-4 Part 4: Edge method
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0

You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum

Search for:
Jump to:  
cron



Hosted by iphpbb3.com

Impressum | Datenschutz